connecting 5890 Chemstation to 6 port valco valve
Discussions about GC and other "gas phase" separation techniques.
8 posts
Page 1 of 1
I am trying to find a 5V TTOL contact closure to operate a 6 port Valco valve on my 5890 operating Chemstation. Is it controlled in Chemstation by the "Valves/Relays" page under Instrument Control? Where can I plug the valve in to get the contact closure?
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
It's a long long time since I have seen a 5890, I do not recall there being a contact closure to operate a valve on the 5890. The valves on the 5890 were air actuated and I remember that there was a special board that had to be installed to enable valves to be controlled. I believe that the very early 5890's only alowed control of two valves, but later models did allow four valves to be controlled. The purge valve function for a capillary inlet was used to control the valves. If you look on the left side of the GC close to the inlets, you should see two sockets close to each inlet position. One is for the heater/sensor and the other is for the purge valve. The control for the valve is either 24 Volt or 12 Volt. You could use this to control a relay which would then give you contact closure for your valve.
Gasman
Gasman
-
- Posts: 4
- Joined: Tue Jan 29, 2013 9:43 am
Thanks gasman. I was lookng at the purge valve and it's 24VDC as you said.
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
If I get a chance today, I'll look at one of our 5890 units, I used it the last two days, and will be using it today as well.
-
- Posts: 4
- Joined: Tue Jan 29, 2013 9:43 am
Thanks KM - I'll fook fed to it.
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
I don't see any electrical contacts on our 5890 units, or software for that, sorry.
-
- Posts: 4
- Joined: Tue Jan 29, 2013 9:43 am
Thanks - I'm going with Gasman's advice above. I'll step down the 24VDC from the purge valve to 5VDC and that should work.
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
Be careful here. I have seen some HP/Agilent 5890 or 6890 where the 24 Volts is supplied to the solenoid valve and the ground side was switched. This meant that there is always 24Volts on the solenoid valve. If the 24 Volt supply is switched, you should be OK.
Gasman
Gasman
8 posts
Page 1 of 1
Who is online
In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 20 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science