SPME Injection
Discussions about GC and other "gas phase" separation techniques.
6 posts
Page 1 of 1
I was wondering if anyone has experience with GC SPME injection. I have gone through 3 fibers in the last two weeks. This seems to happen upon injection into the inlet. I thought that if the syringe was pulled out prior to the inlet purge it would work. Im guessing my inlet flow or pressure is too high or the septa is too thick ???
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
I do it all the time. The fiber has to be in the protective sheath any time it passes through a septa. Just, hit prep run, stick it through the septa, extend the fiber, wait 1 min for desorption, hit start, then retact fiber and pull it out of the inlet. Also what liner are you using. At my last position we did extensive research and found a 1mm Restek 20973 liner with a 2:1 split works best.
The septa purge and purge valve should have no effect on the fiber.
The septa purge and purge valve should have no effect on the fiber.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
If the fiber lost from the asembly? charred? bent?
I will not that a liner with a glass wool plug will remove the fiber from the wire that it should be attached to...
I will not that a liner with a glass wool plug will remove the fiber from the wire that it should be attached to...
-
- Posts: 8
- Joined: Tue Sep 18, 2012 3:18 am
May I add a question to this - what septa and needle do you use? I am having trouble with septa being blown out in just a few injections - using Supelco 24 Ga SPME assemblies with square points.
thanks
Robert
thanks
Robert
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
I am not sure what needle I have. I use the 2cm Carboxen/DVB/PDMS fiber and Agilent green septa. I haven't really had any trouble with SPME at all. In fact I find it very convenient for dry starch encapsulated flavors. The only downside is there is some controversy as the whether SPME provides a representative composition of the sample when doing % reports and the precission isn't as good as liquids injection. However, any technique I sue will have issues especially with volatiles and with different response factors on the TIC.
-
- Posts: 10
- Joined: Tue Mar 16, 2010 8:59 pm
Thank You for time and input on this topic. I will try that approach.
6 posts
Page 1 of 1
Who is online
In total there are 164 users online :: 1 registered, 0 hidden and 163 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 163 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 163 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science