Page 1 of 1
BHA and BHT via GC
Posted: Mon Jan 28, 2013 10:03 pm
by MSCHemist
Does anyone have a good method protocol for doing BHA BHT in edible oils via GC. I am having trouble getting decent recovery especially of BHT. I am dissolving in Pet ether and extracting with acetonitrile but need to do at least 3 extraction to get full recovery of BHT using up a lot of solvent and I don't have a turbovap here to blow it down.
Re: BHA and BHT via GC
Posted: Tue Jan 29, 2013 12:19 am
by KM-USA
See Quantitative determination of BHT in soap products by capillary gas chromatography, » Journal of the American Oil Chemists Society
» Volume 59, Issue 12 , pp 579-582
Works for BHA too, and in fatty acids, fats and oil as well. You won't have the fragrance interferences they had with finished bar soaps.
Posted: Tue Jan 29, 2013 6:34 pm
by MSCHemist
Thanks for the reference but I decided to just dilute the oil .1g to 100ml with ethanol and even at 50:1 I have enough sensitivty for either Scan or SIM but I decided on SIM (why not the S/N is great).
Posted: Mon Feb 04, 2013 3:18 pm
by MSCHemist
One of the issues that just came up is I am able to resolve 2-BHA from 3-BHA. My samples contain ~5% 2-BHA but my refernce standards are all 3-BHA. I quant off the 165 m/z ion which is the base ion for both 2 and 3-BHA but I am concerned the response factors are a bit different so I am hesitant to just add the 2-BHA area to the 3-BHA.
Posted: Sun Feb 10, 2013 11:50 pm
by MSCHemist
Looks like I am almost there. My samples are based in soybean oil the antioxidant is ~20%BHA and 20%BHT and the finished product about 1% BHA and 1% BHT. I started the validation. I am diluting the antioxidant 20000 fold in 96% ethanol (from the liquor store) with di-BHA ITSD and the finished product 1000 fold and doing it in SIM mode with a single quant ion 50:1 split with calibration points at 1, 5, 10, and 25ppm. I am just adding the 2-BHA to the 3-BHA area count.
My recoveries in artificial samples (BHA and BHT dissolved in soybean oil) are 95-105% and the precision is ~2%RSD
There still seems to be a bias about 5-7% low compared to the parent company lab and supplier's COA which use HPLC-DAD.