Advertisement

How to Get Chemstation to Display % M+1, M+2 for ion peaks?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I've been running reverse phase separations on an 1100 HPLC with G1946D detector and Chemstation v. B.04.03.

I haven't figured out yet how to get the percentage of M+1 and M+2 peaks for a given ion peak. Does anyone know how to do this? Does chemstation have the ability to do this?
Incitor Inc.
"Making renewable fuels a reality"
http://www.incitor.com
I'd extract the ion chromatograms, add them to the method in the signal details, and then specify the base peak, +1 and +2 isotope peaks as separate compounds. Since they'll coelute exactly, Chemstation will attempt to add these as three signals relating to the same chemical (i.e. in the table, they'll all have a single # number). You'll have to enter the peaks manually as new compounds, in the same way as you would for an exactly coeluting internal standard. Once you've got the three peaks as separate compounds, you can analyse and tabulate your data as normal, get the whole thing into Excel (batch processing) and then do the percentage calculation there.
2 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry