						Usually HPLC system calibration frequency for GMP is done every 6 months. What are the risks is calibration is done every 12 months or 18 months?
					
									- 
								
 
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
						We have yearly 3rd party qualification.
We do system suitability bracketing our samples each day of cGMP use.
									We do system suitability bracketing our samples each day of cGMP use.
- 
								
 
- Posts: 40
- Joined: Tue Apr 24, 2012 3:36 pm
We have yearly 3rd party qualification.
We do system suitability bracketing our samples each day of cGMP use.
We are in the same boat, although we are doing our first in-house qualification this week.
IMO, the risk of doing yearly vs 6 months is minimal as long as you do system suit for each run. The only change is the size of the investigation if you have a failed qualification, and the potential amount of runs that are impacted.
- 
								
 
- Posts: 377
- Joined: Wed Aug 24, 2011 2:10 pm
						I would suggest 6 monthly is good but if you do the same analyses regularly and can plot control charts from system suit parameters then you have justification for extending your qualification period out to a year.
					
									Where can I buy the kit they use in CSI?
				- 
								
 
- Posts: 2
- Joined: Wed Jan 23, 2013 2:32 pm
Maybe the same criteria could be used to extend qualification period out to 18 months.I would suggest 6 monthly is good but if you do the same analyses regularly and can plot control charts from system suit parameters then you have justification for extending your qualification period out to a year.
- 
								
 
- Posts: 129
- Joined: Thu Oct 04, 2007 7:16 pm
						Regulatory agencies expect you to follow your procedures, it could be 6, 12, or 18  month qualifications.  18 might be too long in my opinion, usually it's 12 months. If you are running samples 24/7 you may need 6 months as I would think the HPLCs would need more frequent preventative maintenace.
					
									- 
																									  
- tom jupille
- Site Admin
- 
								
 
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
						To be a bit pedantic, "calibration" is done every time you run a method. "Performance Qualification" is done to verify the performance of the system. As a minimum, it should be carried after any maintenance work is done on the system, certainly including the regularly scheduled preventive maintenance. Along with most people, I would recommend annual preventive maintenance as a minimum; more frequently might be indicated if there is a high incidence of failures/problems in between. 
If you are regularly running the same methods in a GLP or cGMP environment, control charting of your system suitability results can also give some insight into the need for maintenance and requalification or justification for modifying your SOP to extend the schedule to 12 months instead of 6.
									If you are regularly running the same methods in a GLP or cGMP environment, control charting of your system suitability results can also give some insight into the need for maintenance and requalification or justification for modifying your SOP to extend the schedule to 12 months instead of 6.
-- Tom Jupille 
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
				LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
			7 posts
							Page 1 of 1
					
	Who is online
				In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: Google [Bot] and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 27 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science