reproducible ghost peaks

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
I am having a problem with ghost peaks that appear at 5.9 min and 6.2 min. on all my chromatograms. My mobile phase is acetonitrile, water, 1-pentanesulfonic acid, and acetic acid. I am using a C18 reverse phase column. I have already changed the column, bought all new solvents ( including HPLC grade water) and chemicals, and I still seem to get the ghost peaks. I inject only water and I still see the same two ghost peaks, but if I inject only acetonitrile the ghost peaks don't appear on the chromatogram. I have even tried running my samples on a different LC than the one I usually use. I have no clue what it can be. Does anyone have any ideas.

Need more information:
- detection wavelength?
- is the separation isocratic or gradient? (details if possible).
- are you adjusting pH? if so, are you putting the pH electrode directly into the mobile phase while you adjust?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

The detecting wavelength is 305 nm. The separation is isocratic. The mobile phase is 320 ml of acetonitrile, 680 ml of water, 20 ml of acetic acid, and 0.87 g of 1-pentanesulfonic acid. I combine them and then filter the mobile phase. I don't adjust the pH, but it is approximately 2. The flow rate is 1 ml/min, and each run is 10 minutes. The compound I am intrested in has a retention time of 5.6 minutes. I have done this separation for many months and have never seen the ghost peaks until about a month ago. They appear at 5.9 and 6.2 min and are broad.

The fact that it's an isocratic separation eliminates a lot of possibilities.

It sounds like something coming in with the water. You've already checked the primary suspects. Other possibilities:

- something leaching from your filters? (Try making up the same mobile phase without filtering and see if that helps)

- residual detergent on your glassware? (rinse thoroughly with mobile phase and see if that helps)
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Okay I will try those two possibilities.

Thank you.

I tried not filtering and rinsing thoroughly with the mobile phase and the ghost peaks are still there.

Hi Raquel !!

Are you using only one chanel ?

Have you tried to use the same HPLC for a different analysis ?. It sounds that something is inside your equipment (in injector part) and it eluates each time when you inject .... Have you cleaned the needle and the reodhyne part ?...

Hope it helps !!

Also:
Did you change anything just before the ghost appeared? Maybe a longer injection loop which you no longer fill with sample, thus possibly co-injecting some air?
Or maybe gunk accumulated on the column and injection of water disturbs/displaces this partially? You still get it after you clean the column? Did you inject the water used in preparing the mobile phase? What happens if you inject acetic or pentanesulfonic?
8 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry