Has anyone investigated the difference between headspace vials?
Is it better to go 10 in 20 or 5 in 10?
Also, are screw caps as good as crimped caps?
Also will a 3 mL loop fill give me 3 times bigger peaks than a 1 mL??
My paper in Analytical Chemistry June 1997 shows the effect of the same sample in 20, 10, and 3.35mL vials.

3mL loop fill will not give you 3 times bigger peaks. You will generally be unable to focus the sample and peaks will be 3 times as WIDE rather than as TALL, and resolution will suffer greatly.

You may wish to investigate SPME if you wish to lower your detection limit by 3-5 orders of magnitude.

There are inherent issues of greater variability with SPME, but rather than not detecting a 1 ppb peak, having a 15% RSD of a 1 ppb peak isn't so bad.

If you CAN discuss the particulars of your analysis I would be glad to offer assistance.

best wishes,

Rod
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry