Hexane and Agilent 1946d MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I just had a 1946d Agilent MS installed with an 1100 pump system and DAD detector. When I ran a sample through the column with hexane/IPA as the solvent I got a response on the DAD but the MS was quiet, too quiet.

I'm wondering if maybe hexane doesn't work as a solvent for the MS?
Incitor Inc.
"Making renewable fuels a reality"
http://www.incitor.com
It's not a particularly good choice for electrospray, but what's the sample/analyte?
Here's a useful link on hexane in electrospray: http://web.uvic.ca/~mcindoe/42.pdf
It's not a particularly good choice for electrospray, but what's the sample/analyte?
Here's a useful link on hexane in electrospray: http://web.uvic.ca/~mcindoe/42.pdf
The analyte is chloromethyl furfural, ethyl furoate ester, and some other unknown compounds. We had initially decided to use hexane as a "weak" solvent, as our analytes are more on the polar side. Do you have any suggestions for a good nonpolar solvent for electrospray?
Incitor Inc.
"Making renewable fuels a reality"
http://www.incitor.com
3 posts Page 1 of 1

Who is online

In total there are 71 users online :: 0 registered, 0 hidden and 71 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 71 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry