I have been battling to make my GC-MS work in the CI mode to quantify NitroPAH's. After many trials, finally found the column to be dead, so installed a new column in the hopes of getting good chromatography. But, now after installing the column, the MS has pumped down, but I am not able to get rid of the leaks.
I run a methane Pre-tune and see peaks at 15,17,19,28,30,31,32,40,41,43,44,45,46.
They dont seem to go away. I have cleaned the MS source, checked for any particles (hair, fabric etc..) inside the chamber, grease sealed the side plate, changed the viton-O ring of the vacuum vent valve, tightened the MS interface nut at the GC side, rebuilt the inlet (splitless inlet) at the GC side, checked for leaks in the flow path (upstream from the inlet) for leaks using electronic leak detector (found no leaks there).
Please provide me with some insight so that I have a breakthrough.... this is getting frustrating...
Thanks,
pdixi001.