Page 1 of 2
problem in start up of Varian GC-450
Posted: Wed Jan 23, 2013 7:18 am
by arfeen66
Hello we are using Varian GC-450, from few days when we start the system it takes lot of time to start and some time did start but its lights are blinking without any display,but it start after few switch on & off .
some body help me to sort out this problem
thanks
Re: problem in start up of Varian GC-450
Posted: Mon Feb 04, 2013 10:28 am
by fendersim
This could be a lot of different things.
What firmware is the GC? You can get this info by pressing the "?" button.
Posted: Wed Feb 20, 2013 7:19 am
by arfeen66
software version is 1.08
Posted: Wed Feb 20, 2013 11:44 am
by Johnny Rod
It might be looking for the connection to the PC, have you changed anything? Checked the cables are all in?
Posted: Thu Feb 21, 2013 3:53 am
by arfeen66
we have changed the power supply of PC few days back
Posted: Thu Feb 21, 2013 2:20 pm
by fendersim
Are you running Fixed IP or BootP?
Posted: Fri Feb 22, 2013 9:41 am
by arfeen66
fixed IP
During start up it give following
Lilo 22.8 Loading Bios..........................................
Bios data checked Bypassed
Posted: Mon Mar 18, 2013 7:24 am
by arfeen66
Varian/bruker Technician sort out that CPU board is Faulty (Cp741068 programmed CPU Board)
Please help me for its Price &
We have other same GC with diffrent configuration & same part can we use this part without any trouble with both GC's
Please help me on urgent basis
Advance thanks
Posted: Tue Apr 23, 2013 11:29 am
by martin_a
thanks
Hi, we had the same problem on two instruments (450-GC from 2008, Firmware 1.08). The battery on the mainboard delivered 0.3V instead of 3V, when it was replaced the instruments started up fine.
Posted: Tue Apr 23, 2013 12:11 pm
by Peter Apps
For sure worth a try - a battery has to be a lot cheaper than a new board !
Welcome to the forum !
Posted: Wed Sep 11, 2013 9:33 am
by tjt
thanks
We are also having the same problem on a VARIAN gc-450. How did check the voltage of the battery to know that it was delivering less.
thanks
Posted: Mon Sep 23, 2013 6:24 am
by maher
thanks
Hello
I agree with Martin, one of our customer had same problem (the unit does not complete booting and you get blinking cursor), there is a battery on the main board looks like coin, I cannot remember part number), the easy way remove this battery and look for replacements locally(2 or 3 $ price), replace it and it should work.
Good luck
Posted: Fri Dec 06, 2013 12:43 pm
by alex067
Same problem today, same solution:
cycling LED colors on the panel, Linux doesn't start, so I changed the battery CR2032 on the 450-GC motherboard, and everything just work like a charm again :D
It was the same problem with older versions too (Varian 3400), and often with PC-s (forget the time, old files cause problems with the virus scanner etc.).
Cheers, and thanks for the forum!
Alex
Posted: Wed Feb 12, 2014 2:02 am
by pradeep163
We have a Bruker 450-GC and experienced the same main board issue. Can some one tell me how to locate the main board battery in 450 GC system?
Posted: Wed Oct 16, 2019 7:34 pm
by Studysoils
I am having similar problems with the Varian 450GC--- screen on but nothing on it and all LEDs blinking in order. It could be the battery on motherboard....what type and kind of battery should I look for when troubleshooting? Thanks for the help.
