High pressure and quick aging in Metrosep IC columns
Posted: Tue Jan 22, 2013 11:01 pm
Hello Everybody!
I’m just getting started on the forum
We have the Metrohm-Applikon’s Monitor for Aerosols and Gases in Air (MARGA). The current matter is about the IC columns’ pressure. The instrument uses an Anion A Supp 10-75/4 column (Polystyrene-divinylbenzene copolymer with quaternary ammonium groups) and a Cation C 4 100/4 column (Silica gel with carboxyl groups).
As soon as we install the columns, the pressure starts rising but there’s also a quick aging of the stationary phase, since the RTs get smaller and the peak quality decreases. The current anion columns are lasting about 7 days, meanwhile the cation columns 20 days. According to the operation manual, they should last 2 months for the anion column and 4 months for cation column.
We are using tridistilled water with conductivity lower than 1 uS/cm, we have done lots of system cleaning procedures with H2O2, and even if we change all the filters and columns, the pressure keeps rising.
I would really appreciate all your comments regards this, I feel I have proved everything and nothing seems to work…
Thank you so much for you attention. Best regards & much success! Angelica