ppm nomenclature is so very ambiguous ! wt/v, wt/wt, v/v, etc. One should clear scientific nomenclature. SORRY to be so snotty, but I know many will agree with me about this. So I am guessing, correct me if I misjudge:
5 microliters of a 1000 ppm solution of toluene in DMAc, or 5mcL of 1mL of toluene per liter of DMAc, or 5mcL of 1mcL per mL of toluene in DMAc, or 5 mcL of a a 1 nL toluene per microliter of DMAc.
That means you are adding 5nL of Toluene per 5mL of unknown sample = 1 ppm v/v
OR
4.32ng per 5mL of unkown sample
Am I right so far?
If the sample is aqueous then 10 min is enough heating, if non-polar, then may be 15 min.
I would expect at least a less than 3% RSD. A simple vial of water can test your sample vent line for bubbles. When your sample stops bubbling through or if your sample loop has been flushed 3-5 times, then that is enough time, but the sample flow MUST be stopped and the line come to equilibration before you inject.
best wishes,
Rod