Agilent 1200 negative peak

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I am trying to detect three compounds which absorb at 488, 472 and 470. The original aqusition method called for acetonitrile but we are now trying to switch to methanol. With acetonitrile, there was negative abosorption before the compounds eluted and a large negative peak after (solvent peak). We have switched to methanol as our solvent B and I am having issues with getting the baseline to be nice and steady. I am also having some sample media (XAD-7 tubes) interference which I'm hoping to cure with any suggestions you may have as to how I can zero out my baseline. Our refernce spectrum is 474 and our range is from 440 to 500, negative absorption is 100.
Correction our signals are set up for, Sample 474 Bw 4, Reference 474, Bw 100. The range is 440 to 550 not 440 to 500.
Turn off your reference wavelength, or set it to a region where your analytes do not absorb.

http://www.chem.agilent.com/Library/Sup ... faq184.pdf
Time flies like an arrow. Fruit flies like a banana.
3 posts Page 1 of 1

Who is online

In total there are 40 users online :: 5 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Bing [Bot], Google [Bot], Google Adsense [Bot] and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry