effect of sample concentration

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
I just want to have simple question. I'm running my venlafaxine sample via HPLC using 250X4.6mm, 5microns particle size phenyl column. What would be the effect of changes in concentration on its retention time. Let say between 0.001mg/mL against 1mg/mL. Thanks. by the way, I'm using gradient condition where my mobile phasse are buffer pH 3.0 and 100% Acetonitrile.
So long as you are not overloading, retention time should be independent of concentration. At high levels, overloading *usually* results in decreased retention time and increased tailing. Occasionally (particularly if your analyte has limited solubility), retention time will increase and you will see fronting.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
So long as you are not overloading, retention time should be independent of concentration. At high levels, overloading *usually* results in decreased retention time and increased tailing. Occasionally (particularly if your analyte has limited solubility), retention time will increase and you will see fronting.
originally, i prepared 1 mg/ml concentration of sample, afterwards i injected it and I was getting approximately 6.985 minutes retention time. From the same 1mg/mL sample, I diluted it to have 0.001 mg/mL and when I injected it, i am getting 7.3 minutes as retention time. Why on this case, the results seems dependent with concentration? As from my method, it was noted that 1mg/ml sample retention time should be having +/- 0.1 difference only from the retention time of 0.001mg/mL but base from what I'm getting, results did not follow the premise given. Any possible reason behind?
I'd make a linearity experiment, and inject concentrations from your lowest to highest (same injection volume), and then inject lowest at the end. If your retention times are less for the higher concentrations, likely would be that there's more molecules present for a finite amount of column particles/phase.
results did not follow the premise given. Any possible reason behind?


What you are seeing is perfectly reasonable: at 1 mg/mL, retention time was shorter than at 0.001 mg/mL, which implies that at 1 mg/mL you are seeing some overload effect. CPG's advice is spot-on: run a series of calibrators to establish the linear range.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
CPG's advice is spot-on: run a series of calibrators to establish the linear range.
Thanks, Tom. Logical is 98% of the game in my estimation.
In my opinion, the logical approach is way better than the lethargic one.
7 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry