Agilent 1100 injector problems
Posted: Mon Jan 21, 2013 8:52 pm
When my system's injector is on mainpass mode, there is a leak on the needle seat and the pressure fluctuates during the injection process and the run pressures are low. Peaks are broad and injection spike occurs at 6 minutes.
On bypass mode, the leak stops with normal run pressure and peak shape. When the instrument runs with ADVR checked, the problems are resolved. I just installed a new needle seat, and I am still seeing the problems. I looked at the rotor/stator and they appear to be intact and clean. The system was PM’d 5 months ago with a new rotor and stator. We are getting by with ADVR mode; however, I’m concerned that we will have problems with carryover in the long term.
Can someone offer a suggestion to fix the problem or further troubleshoot this?
Thanks,
Denise
On bypass mode, the leak stops with normal run pressure and peak shape. When the instrument runs with ADVR checked, the problems are resolved. I just installed a new needle seat, and I am still seeing the problems. I looked at the rotor/stator and they appear to be intact and clean. The system was PM’d 5 months ago with a new rotor and stator. We are getting by with ADVR mode; however, I’m concerned that we will have problems with carryover in the long term.