bh.buaa » Thu Jan 24, 2013 4:06 pm
We have a Agilent GC6890, and yesterday we tried to calibrate the capillary column with FID detector on GC. The calibration gas components are C2H2, C2H4 and C2H6. The carrier gas in the GC and balance gas in the calibration gas are Helium. When we calibrated it, a issue happened that the FID flame kept switching on and off in the calibrating process until the first peak of C2H2 generated. We calibrated the GC with the some gas sample four time, while the FID flame was keeping on only at the first time. In the next three calibrations, the FID flame just switched on and off as I mentioned above. Also the times of flame on and off for different calibrations looks increase. When the flame on, there was also a sound like explosion.
Again, the flow rate for H2, Air, makeup gas and column are 30ml/min, 300ml/min, 26ml/min and 4ml/min. The temperature of the FID detector is kept on 300 degree C.
Is anyone met this issue before? What is the reason for this issue? How to solve it? Thanks a lot.