we're setting up a GC with a FID and we are getting very high FID signal values. We get these results:
1- [Zero air (generator) + trap] + [H2 (cylinder) + trap] --> 260 pA
2- [Zero air (generator) + trap] + [H2 (cylinder)] --> 67 pA
3- [Zero air (cylinder) + trap] + [H2 (cylinder)] --> 65 pA
4- [Zero air (cylinder) + trap] + [H2 (cylinder) + trap] --> 267 pA
It seems like the trap from the H2 (Molsieve 13X) source is making the difference.
What would you recommend us to do? Should we leave the GC for a long period of time and see if this value changes? We left it for one night with configuration #1 and the signal diminished a little (from 330 to 260 pA)...
Thanks in advance for your help!