Automatically turning Agilent 1100/1200 off after sequnce

Posted: Sat Jan 12, 2013 12:19 am
by CLSgroup
This is probably something very simple but I cant seem to find it in the manual. How do I set up my aqusition method to end after my run has ended. Do you have to create a special stop method? Using Agilent 1200.

Re: Automatically turning Agilent 1100/1200 off after sequnc

Posted: Sat Jan 12, 2013 7:13 pm
by akthmps
Under "sequence parameters", you can select for a macros to automatically run at the end of your sequence. Select the "shutdown" macros to turn off the components of your system.

Re: Automatically turning Agilent 1100/1200 off after sequnc

Posted: Sun Jan 13, 2013 5:19 am
by Consumer Products Guy
Under "sequence parameters", you can select for a macros to automatically run at the end of your sequence. Select the "shutdown" macros to turn off the components of your system.
Yeah. Choose Standby to shut off all, or PUMPall, or LAMPall.