MS library
Posted: Fri Jan 11, 2013 6:41 am
by ito_miura
Hi All,
Does anyone use MS library except NIST that give quickly for identification every peak?
Thank's for the info ...
Ito
Re: MS library
Posted: Fri Jan 11, 2013 9:21 am
by Peter Apps
As far as I know it is the MS software,not the library, that determines whether you can just run through a chromatogram and generate a list of best fit identities for all the detected peaks.
Re: MS library
Posted: Fri Jan 11, 2013 3:30 pm
by Consumer Products Guy
We have NIST library, but also have generated our own library of consumer products ingredients and fragrance components.
Re: MS library
Posted: Sat Jan 12, 2013 12:55 am
by West Chemist
We also use Wiley's library. Wiley is great as well. There are certain compounds that Wiley will find for us that Nist won't. But for the most part they agree well with one another and are pretty comparable to each other. But at times it can be nice to see the little details that are different between Nist and Wiley so as to compare to the spectra from my instrument when making judgement calls.
In addition we have built our own library which since it was built on our own runs of standards, of compounds we expect to potentially see, we usually will get quite nice matches there.
Depending on the library your software has (could have all but five compounds in the library) or something like NIST with tens of thousands of compounds you will certainly stand to see more spectra matching to the one you produce with the larger library. But sure the software needs the capability. And I know from the types of compounds we test for, there often are still compounds that go unknown.
Re: MS library
Posted: Sat Jan 12, 2013 3:18 am
by Don_Hilton
The phrase "identification for every peak" stands out to me. Let me repeat one thing that Peter pointed out: best fit identifies.
Some people expect mass spectral libraries to provide correct compund identificaiton to every peak in a chromatogram. While this may happen, it often does not. There are several places on the site where the reasons for this are disucssed.
Re: MS library
Posted: Mon Jan 14, 2013 3:13 pm
by lmh
yes, that phrase scares me. A library as big as NIST will identify every peak in the sense that every peak will provide a hit of some sort, but depending on your sample, it is highly likely that many peaks are wrongly identified.
This isn't the software's fault, or NIST's. It's simply that there are more chemicals in the world than can possibly be put in one library (at the moment) and the software is only supposed to find the best fit, not check whether it's definitely right. Some things, in any case, are extremely difficult to distinguish by mass spec alone.
Re: MS library
Posted: Tue Jan 15, 2013 3:34 am
by MSCHemist
I have NIST, WIley, Flavors, and custom additions and still frequently get peaks with no match.
Re: MS library
Posted: Tue Jan 15, 2013 7:53 am
by ito_miura
Thanks All for clearing our misconceptions on peak identification
As Peter said, it is the MS software,not the library that determines whether onecan just run through a chromatogram and generate a list of best fit identities for all the detected peaks. We have Perkin Elmer, Clarus 600 T with TurboMass software. Must we use this software or can we use other software with our MS? Because it is very time consuming to identify major peaks one by one.
Re: MS library
Posted: Tue Jan 15, 2013 8:52 am
by Peter Apps
I would be surprised if you could run your MS using other software, and even more surprised if the software that you have cannot do automated library searches if you set them up in the data processing part of a method.
Re: MS library
Posted: Tue Jan 15, 2013 9:52 am
by Jetjamnong
In the turbomass SW, you can change searching parameter to obtain more faster result within 2-3 second or you can open up with NIST/AMDIS and do library search. An another way to do library search using turbomass is by selecting a communique' report in the sample sequence table at the qulitative analysis column and setup this, its can provide you all matching results within a minute depend on your criteria.