Strange peak in LC/MS
Posted: Thu Jan 10, 2013 10:13 am
Hi there!
I've conducted my senior internship in the Academic Hospital of Maastricht Netherlands. My project is about the identification of hemoglobin variants with LC/MS or LC/MS/MS. Everything went good so far but since November 29th I've a problem.
When I measure blanks, I get a strange peak which elutes at 15 minutes. When I analyze this peak, I mainly receive two m/z values which are 253.1 and 505.0.
HPLC: HP 1100 Series
Column: Zorbax C-18 at 35 Celcius with a C-8 guard column
Flow: 0.2 ml/min
Mobile phase: A - 2mM Ammoniumacetate 0.1% FA
B - Acetonitril 0.1% FA
Gradient: 0-15 min 40%A 60%B
15-16 min 40%A 60%B
16-17 min 98%A 2%B
17-25 min 98%A 2%B
MS: Agilent 6410 QQQ on full-scan because I'm still optimizing my method
It seems that the problem is the MQ water and I changed all mobile phases to ULC/MS water and the peak disappears but now it's back again..
Anyone suggestions
I've conducted my senior internship in the Academic Hospital of Maastricht Netherlands. My project is about the identification of hemoglobin variants with LC/MS or LC/MS/MS. Everything went good so far but since November 29th I've a problem.
When I measure blanks, I get a strange peak which elutes at 15 minutes. When I analyze this peak, I mainly receive two m/z values which are 253.1 and 505.0.
HPLC: HP 1100 Series
Column: Zorbax C-18 at 35 Celcius with a C-8 guard column
Flow: 0.2 ml/min
Mobile phase: A - 2mM Ammoniumacetate 0.1% FA
B - Acetonitril 0.1% FA
Gradient: 0-15 min 40%A 60%B
15-16 min 40%A 60%B
16-17 min 98%A 2%B
17-25 min 98%A 2%B
MS: Agilent 6410 QQQ on full-scan because I'm still optimizing my method
It seems that the problem is the MQ water and I changed all mobile phases to ULC/MS water and the peak disappears but now it's back again..
Anyone suggestions