Strange GC issue
Posted: Wed Jan 09, 2013 5:39 am
Over the last couple of months I have started to do quantitation on timber samples and have been having an interesting time with some uncertainties with one of the analytes.
For the most part I have two actives in methanol to which a DBP internal standard is added. The issue I am having is one of the analytes is giving consistently varied results even from the same sample. The other analyte is fine however. Also RRFs are reasonable..
I have found that some of the troubled analyte is being hung up pre-coloumn as a flush after the standard shows. The amount that is being hung up is varied as well. The pre-coloumn is tapered with glass wool packing which, when changed stops this 'sticking' of the problem analyte for a short while before the problem returns and slowly grows over time.
To add to this, when I do this whole procedure with acetone instead of methanol the problem is gone. This is with a split-less injection of which there is far too little analytes in the standard or sample to overload the coloumn.
Any ideas on what is causing this issue?
For the most part I have two actives in methanol to which a DBP internal standard is added. The issue I am having is one of the analytes is giving consistently varied results even from the same sample. The other analyte is fine however. Also RRFs are reasonable..
I have found that some of the troubled analyte is being hung up pre-coloumn as a flush after the standard shows. The amount that is being hung up is varied as well. The pre-coloumn is tapered with glass wool packing which, when changed stops this 'sticking' of the problem analyte for a short while before the problem returns and slowly grows over time.
To add to this, when I do this whole procedure with acetone instead of methanol the problem is gone. This is with a split-less injection of which there is far too little analytes in the standard or sample to overload the coloumn.
Any ideas on what is causing this issue?