Hydrogen as GCMS carrier gas.
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
Anybody use hydrogen? I did a quick search and didnt see this topic. I'm sure its been brought up several times, but couldnt find it. Perhaps because the folks who tried it all went boom? As helium is a finite resource and rationing is occurring, the economical move would be to use hydrogen. Does anyone have any safety tips on using hydrogen as a GCMS carrier?
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
This is a hot topic right now. Agilent has some info for their new instruments on their website. Included in that is the safety warnings and features. Other vendors have information too. Thermo has a webinar out there somewhere, sorry I deleted my link, but a quick google search should find quite a bit. EPA's Offices of Solid Waste and Water are looking at what they may need to do in the future for their methods.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
I am currently working on trying the transition myself. Still having trouble with volatiles passing a BFB tune, but made the changes Agilent recommended and trying some more. The only problem so far has been getting the m/z 96 to m/z 95 ratio below 9%. Will be trying it soon on semivolatiles with the DFTPP tune criteria. No problems with the volatiles so far except this.
The Thermo webinar said they had to bake out the source with a high flow of H2 overnight, something like 4ml/min at 350c to get the ratio to fall into a passing range.
I had 96 down to 16% last week but cleaned the source and switched to the larger diameter drawout plate and went to the 0.18mm column and now it is back up to 50%. Previously it started at 80% and dropped to 16% after a week of running test samples. Something bakes out over time that is causing the problem I hope we can get it into passing range so we can make the switch.
If you are not under EPA tuning criteria then it probably would not be such a bad transition as long as you look at new ion ratios for your targets, just need to run standards and update the ratios. The ion gauge does read higher than with Helium but as long as you keep the reading below 6*10^-5 torr you should be ok with flow.
The Thermo webinar said they had to bake out the source with a high flow of H2 overnight, something like 4ml/min at 350c to get the ratio to fall into a passing range.
I had 96 down to 16% last week but cleaned the source and switched to the larger diameter drawout plate and went to the 0.18mm column and now it is back up to 50%. Previously it started at 80% and dropped to 16% after a week of running test samples. Something bakes out over time that is causing the problem I hope we can get it into passing range so we can make the switch.
If you are not under EPA tuning criteria then it probably would not be such a bad transition as long as you look at new ion ratios for your targets, just need to run standards and update the ratios. The ion gauge does read higher than with Helium but as long as you keep the reading below 6*10^-5 torr you should be ok with flow.
The past is there to guide us into the future, not to dwell in.
3 posts
Page 1 of 1
Who is online
In total there are 196 users online :: 1 registered, 0 hidden and 195 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 195 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 195 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science