We've been running a stability-indicating SEC-HPLC method on Agilent 1100 for the past 30 weeks or so; the samples are stored at 2-8C and appear to be fairly stable. I compared the total peak area (280nm) from different timepoints and found it to increase in a smooth function over time. For example, at week 0, the total area was 4000 mAU*s, whereas at week 30, it was 4400 mAU*s. Filling in timepoints in between (e.g., 3, 6, 12, and 20 weeks) shows that the increase in total peak area is somewhat asymptotic, so that a plateau appears to be nearly reached.
This might be a problem with recovery from the column, but I have tried a couple of different columns at the week 30 timepoint and found the total peak area to be ~4400 mAU*s. Therefore, I don't think it's a column issue.
I also don't think it's a sample issue because I also run samples stored at different temperatures (-70C, -20C as stable controls) and they all show the same smooth increase in total peak area.
Could this observation be due to lamp life? Can one unequivocally say that the older the lamp, the greater the peak area (assuming the sample remains unchanged)?