HEEEELP! Gradient bad shape!
Posted: Wed Jan 02, 2013 12:02 am
Dear all,
I am struggling since two weeks with my HPLC. I am running a gradient with the following conditions:
0-5 min: Isocratic at 40%ACN/Water
5-25 min: Linear Gradient 40% ACN/Water to 100% ACN
25-35 min: Isocratic at 100%ACN
35-45 min: Isocratic at 40%ACN (Re-Equilibration)
Low-pressure mixing pump, Flow = 1.5 ml/min, UV Detector at 254nm, Temp.=30 oC, C18 column 250x4.6mm 5um with guard column, online degasser, autoinjector set at 20 ul injection. Sample solvent is ACN and samples have low levels of analytes.
The PROBLEM is:
I used to get a good (I think!) linear gradient shape with the two isocratic plateaus at the start and end of the gradient with a maximum height around 0.0005Volts. However, after running a number of samples the shape started to deteriorate with a peak (or a sudden jump) at about 17 min, and absorbance levels increased.
Below is a chromatogram with the BAD shape:
In addition to the BAD shape, the absorbance levels are increased gradually from an injection to the following one!
My investigations: ( ALL INJECTIONS WERE ACETONITRILE)
1st : I washed the column with 100%ACN for 30min, then repeated 3 injections and got the chromatogram below:
Better shape and lower absorbance! But there was still some deterioration in the shape and the absorbance levels are still increased gradually from replicate 1 to 3 and deteriorated more with the following injections.
2nd : I consulted this forum and watched the Mini-tutorial on gradient artifact peaks posted thankfully by Tom Jupille. I then ran the "three blank gradients" test. The results are shown below:
I think there was no clear evidence of contamination of the A solvent (Water), so I continued having fun! I suspected that something is happening from the injections.
3rd : I injected ACN 10ul, 20ul, 60ul, 100ul and then 10ul again to check if the sample solvent is causing the contamination. However, there was no clear trend of the absorbance with the injected volume.
4th : I suspected the Injector so I replaced the Injector wash and washed the Injector with 25 injections of 100ul ACN. Then ran 4 replicate injections of ACN. The results are shown below:
I think things are becoming worse! The absorbance level increased more and the shape deteriorated more!
5th : now it is the column time! I disconnected the column and replaced it with a tubing to get some backpressure and then ran 2 injections. I got good shape and absorbance: linear gradient shape with the two isocratic plateaus, and a maximum absorbance level at 0.0007V. So, I suspected the column has accumulated some garbage from somewhere.
6th : I disconnected the guard column and made 2 injections to get the below:
The absorbance level dropped significantly BUT the shape is still not OK and the absorbance increased from the first to the second injection!
7th : I tried a another column (a used C18, 150x4.6mm, 5u) BUT I got almost the same shape and level of absorbance.
8th: I back to my column (the 250mm) and washed the column and the system with 80%ACN at 0.5ml/min for about 20 hours!!! I got the below:
Better shape and absorbance levels but the increasing levels of absorbance between injections is still there!
9th: I stopped having fun and started to cry because I wasted most of the ACN purchased for the current project!!!
So, I decided to consult you in this valuable forum through this post!