Calibration Procedure for Agilent 7820A GC

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hello
We have recently purchased Agilent 7820A GC with 7697A Headspace sampler.
We have only 1 FID and Single injector port with transfer line.
We are using this instrument for Residual solvent test as per USP.
As we are using this instrument for pharma, my question is How to calibrate this instrument?
I am using BP624 capillary column for analysis purpose.

Thank You.
One does not calibrate the instrument, rather you process standards which allow you to verify the reproducibility and level of detection of your analysis for any solvent using headspace.

The proper operation of the instrument is also verified by using standards which confirm the sensitivity and reproducibility of the detector.

The procedure for this confirmation of proper operation of the GC is found in the installation documentation of the instrument.

Good luck.

Rod
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry