Dear users,
I have a method using copper in the mobile phase and was looking for guidance on flushing the system to make sure all of the copper has been removed from the system. Could I please get input on the best method/mobile phase to flush the system with?
Regards,
Mark
I'd be inclined to pull the column and flush w/ some 2-3N HNO3, followed by lots of water.
Thanks,
DR
Image
I assume you are using a chiral column... and I'm probably running a similar one...

What we do is take the column off, flush the system well with water and then with organic in preparation for a different column/mobile phase...

We don't see an issue... What is the problem you see?

- Karen
I assume you are using a chiral column... and I'm probably running a similar one...

What we do is take the column off, flush the system well with water and then with organic in preparation for a different column/mobile phase...

We don't see an issue... What is the problem you see?

- Karen
Karen,
I haven't seen an issue yet. Last time I tried the nitric acid flush suggested by another reader here, but it took several hours to flush all of the nitric from the system so I was looking for an alternate flush procedure to make sure the copper was gone. I am actually determineing EDTA and DTPA on a RP column and have nightmares about the copper carrying over into another column and analysis.
Thanks for your reply,
Regards,
Mark
For EDTA: try HAMILTON PRP-X100, 4.1 x 150mm 10μm, 250 nm, pH:3.5 buffer:acetonitrile (75 : 25) after complexation of EDTA with ferric chloride..

Good luck
I am actually determineing EDTA and DTPA on a RP column and have nightmares about the copper carrying over into another column and analysis.

Mark, we do EDTA and DTPA on a C8 column using copper in the mobile phase, and don't see any real issues. I feel for this that it's best to dedicate a column for this assay, but that may be more because the mobile phase we use is aqueous.

We flush the system and column with HPLC H2O for 15 minutes at 2 ml/min., then pump through with about 20% HPLC methanol or 20% acetonitrile.
I am actually determineing EDTA and DTPA on a RP column and have nightmares about the copper carrying over into another column and analysis.

Mark, we do EDTA and DTPA on a C8 column using copper in the mobile phase, and don't see any real issues. I feel for this that it's best to dedicate a column for this assay, but that may be more because the mobile phase we use is aqueous.

We flush the system and column with HPLC H2O for 15 minutes at 2 ml/min., then pump through with about 20% HPLC methanol or 20% acetonitrile.
Would it be possible for you to share your method for the EDTA/DTPA? Will send you my e-mail address if it is OK for you to share.
Would it be possible for you to share your method for the EDTA/DTPA?
Unfortunately, that is considered proprietary, from a supplier of chelating chemicals. All I can share is that the mobile phase is a mixture of quaternary ammonium surfactant and copper sulfate at pH 4.2 with acetic acid/NaOH (no organic solvent) and samples are made up in dilute copper sulfate. RP-8 and UV detection is used.
8 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry