Hello,

I have several samples measured using HPLC-DAD with a gradient. My problem is the reference wavelength I have chosen to zero the baseline might not be optimal, I've been getting negative spectra consistently. Is it possible to recover the original signal, without baseline substraction after the sample has been measured? or is it possible to change the reference wavelength after measurement?

I am using an Agilent chromatographer and ChemStation software.

Thank you.
Dear damt787,

I have looked for the same feature and have not found it. When I've run into the same problem I recollect with the reference off & using a wide range on my UV acquisition (190-600nm+), check the UV spectra of my peak, and evaluate whether to use a different reference or keep it off based on what the spectra tells me.

Sincerely,
Keith
2 posts Page 1 of 1

Who is online

In total there are 39 users online :: 2 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry