Hi and thanks in advance for helping out.
I am using acetic acid adjusted to pH 5 with ammonium hydroxide on A and 100% MeOH on B and I am monitoring at 230nm using a phenyl column. My problem is that the baseline slopes downward and I would like to correct for it and don't know how, any help would be appreciated.
your going to get slope when using a gradient you really cant help that. . You could try to raise your wavelenght higher to a region where there is less absorption but you may lose sensitivity or you could go isocratic
Hi Shadow,

Acetate and 230 nm detection is a bad combination. Use another acid/salt instead of acetic acid/acetate.
What is the pH in you mobile phase by the way?

Best Regards
Learn Innovate and Share

Dancho Dikov
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry