220V GC/MS anything I need to be concerned about
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
11 posts
Page 1 of 1
Hi all. I am ready to try again on getting a refurb 6890-5973. This time I am going with a reputable refurb company. The only thing is it is a 220V system. I know I will need to get a 220V rough pump if I want to plug it into the MSD. Are there any other concerns. Can it plug into a standard outlet?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I believe there are several types of 220 VAC plugs and receptacles, you'll need your electrician or someone to make sure to match up to a correct receptacle. So you might need to swap out the plug or the receptacle, I'd wait until it arrives on site. Then you might need a quick trip to Lowe's, H Depot, or Ace, unless you have an electrician on-site.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
If the system is a 220 V system, you will need a 220 V outlet for the system. If the system is originally from Europe, you may even need 220 V outlets for the computer and accesories. Get the installation guide. Typically in the US a "220 V" outet is acutually 208 V (but, if I reall correctly, measure 220 peak to peak). Some 220 V systems can run on 209 V while others require a step up transformer to bring the ine voltage up to 220 V (RMS).
-
- Posts: 196
- Joined: Tue Oct 04, 2011 11:02 pm
Follow Don's advice. A lot of the older instruments required 220 for the MS. You do need a buck/boost transformer to bring the normal 208 volts up to 220 because of the turbo or diffusion pumps that are running off the internal power. As far as the much pump is concerned, most have internal jumpers so you can run them on 110 or 220. If you decide to plug it into the wall outlet, make sure it's on the same circuit the MS is on. If it isn't and you lose power to the pump and not the MS, you will back stream.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
The computer we are providing. Do the Edwards E2M1.5's have the internal jumpers? Agilent has 2 different part numbers for both Pfiefer Duo's and Edwards E2M1.5's one for 110 and one for 220.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
The electrician want to know if it is single phase or three phase?
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
I found the siteprep checklist for the 5973. It seems to indicate that 208V-WYE is perfectly acceptable on page 25.
http://www.chem.agilent.com/Library/use ... 035252.pdf
http://www.chem.agilent.com/Library/use ... 035252.pdf
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
Interesting according to the preplist the GC and MSD itself can run on 120 or 220/208 without reconfiguration. However, the diff and foreline pump have to match the voltage range that is being supplied to the MSD. It looks like the GC can run on either as well though not sure about the 7673/g1512a ALS.
It looks like the exception is a fast heat 6890 oven 208 can't work for it. Hmm according to the 6890 manual
http://www.chem.agilent.com/Library/Sup ... a16012.pdf
it needs to be specific to that particular GC.
feh it's all so bloody confusing.
It looks like the exception is a fast heat 6890 oven 208 can't work for it. Hmm according to the 6890 manual
http://www.chem.agilent.com/Library/Sup ... a16012.pdf
it needs to be specific to that particular GC.
feh it's all so bloody confusing.
-
- Posts: 50
- Joined: Sun Nov 06, 2011 6:55 pm
Many European 220V requirements are single phase to Neutral, as opposed to US where you tend to find 220V (or 240V) measured peak phase to phase. This in not your home washer/dryer setup (2 x 120V) rather a true (usually 13-15A) single phase 220 (or 240V) setup. Have your house electrian (or hire a licensed one) to check and match to installation requirements.
-
- Posts: 310
- Joined: Thu Jun 18, 2009 7:59 pm
If you have the 208 available you definitely want to run the GC on it - you'll get substantially faster heating rates. Since it's a '73, the pump is small and won't make a lot of difference, but the GC will pull fewer amps (which is actually what you pay for).
It's all single phase. Make sure you are running off of a single phase box, and not a three phase. If you're running off a 3 phase box make sure you get the phases correct when the plug is wired!! Many electricians do not know how to do this.
Also, if you can, upgrade the E2M1.5 - it's just too small. Get a real rough pump.
It's all single phase. Make sure you are running off of a single phase box, and not a three phase. If you're running off a 3 phase box make sure you get the phases correct when the plug is wired!! Many electricians do not know how to do this.
Also, if you can, upgrade the E2M1.5 - it's just too small. Get a real rough pump.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Laboratory Director
Krause Analytical
Austin, TX USA
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
Yep Gentech recommended an E2M2 so that's what i'm going with.
I thought it might be too powerful to plug into the back of the MSD so I just got an E2M1.5 for my 5975. It seems to be identical to the Pfiefer Duo (that I hate) in terms of the final vacuum reading and how long it takes to pump down.
I thought it might be too powerful to plug into the back of the MSD so I just got an E2M1.5 for my 5975. It seems to be identical to the Pfiefer Duo (that I hate) in terms of the final vacuum reading and how long it takes to pump down.
11 posts
Page 1 of 1
Who is online
In total there are 44 users online :: 1 registered, 0 hidden and 43 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 43 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 43 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science