UPLC reliability issues
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
Ive had reliability issues with our waters UPLC for some time now. I have posted multiple times about building pressure issues and the only way i solve it is by taking the head off the column and replacing the 0.2um filter. Thta works for a while until i need to do it again. Waters have blamed just about everybody and everything...filters...technique...solvents and bascially i believe i have eliminated all of these. I have no idea what cause this problem!!...any ideas...cause im totally out of them!
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
What makes you believe that you have eliminated all possible sources of whatever is clogging the filter ?
Peter
Peter
Peter Apps
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
When I was running my Agilent 1290 to avoid building pressure I'd
1. Not use pure water on any channel. I used 10% CH3CN on channel A
2. I always ended a run with a cleaning that left the column in 100% CH3CN
3. I used an inline filter. the smallest available for Agilent was 0.3um
4. I used a bizarre but effective cleaning procedure where I alternated between A water with 0.1% TFA acid and B 33% IPA/66%CH3CN/0.1%TFA. I ran a 30min gradient from 20% to 100%B at 80deg (max column tolerance) 3 times with max volume injections of trifluouroethanol. I did a lot of experimentation with cleaning columns and this one was a miracle. I even tried running heptanes at low flow overnight to clean the column and that usually didn't do much (I ran reverse phase)
1. Not use pure water on any channel. I used 10% CH3CN on channel A
2. I always ended a run with a cleaning that left the column in 100% CH3CN
3. I used an inline filter. the smallest available for Agilent was 0.3um
4. I used a bizarre but effective cleaning procedure where I alternated between A water with 0.1% TFA acid and B 33% IPA/66%CH3CN/0.1%TFA. I ran a 30min gradient from 20% to 100%B at 80deg (max column tolerance) 3 times with max volume injections of trifluouroethanol. I did a lot of experimentation with cleaning columns and this one was a miracle. I even tried running heptanes at low flow overnight to clean the column and that usually didn't do much (I ran reverse phase)
3 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science