Liquid chromatography was carried out with a Agilent HPLC-System (Series 1200). The LC column is a silica-bonded ZIC-HILIC, 2,1 x 150 mm, 5 micro-meter.
Solvent A is 250 ml H2O, 250 ml ACN and 0.38 g NH4Ac
Solvent B is 25 ml H2O, 475 ml ACN, 25 micro-liter HCOOH
mobile phase gradient
time A B Flow
0.0 0 100 0,4 ml
3.5 0 100 -"-
6.0 35 65 -"-
6.5 35 65 -"-
7.0 75 25 -"-
9.0 75 25 -"-
9.2 0 100 0.6 ml
12.0 0 100 0.6 ml
12.5 0 100 0.4 ml
18.0 0 100 0.4 ml
Mass spectrometry was carried out with a Applied Biosystems (API 4000).
My problem is that I haven't any experience in the work with a ZIC-HILIC column. The retention time for melamine changes from run to run. From 8,35 min at the beginning till 5,49 min after serveral runs (may be 30 runs). After approximately 30 runs, the retention time is stable. The retention time for cyanuric acid is stable from the beginning.
I already tried an other gradient from the US-FDA. But every day the same problems.
Any suggestion? What could be the problem?