Anybody know how to measure glycerol? I thought ion exlusion but I know glycerol is not stable in acid.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
RI detector with 100% water mobile phase on a C18 works wonders.
best wishes,
Rod
best wishes,
Rod
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
We've used capillary GC with BSTFA derivatization to do this for decades, see "Quantitative Determination of Glycerin in Soap by Capillary Gas Chromatography", JAOCS, Vol. 64, No. 9 (September 1987)
We use this also for liquid products, and for similar materials such as propylene glycol, glycol ethers, sorbitol, etc. Very straightforward and easy.
We use this also for liquid products, and for similar materials such as propylene glycol, glycol ethers, sorbitol, etc. Very straightforward and easy.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
CPG
He is doing biodiesel waste which should contain a bunch of water.
He may wish to make tri-acetates by GC or simply run a HPLC method which if he has one makes a lot of sense in a production mode.
It depends upon the actual content of his matrix which can vary.
Rod
He is doing biodiesel waste which should contain a bunch of water.
He may wish to make tri-acetates by GC or simply run a HPLC method which if he has one makes a lot of sense in a production mode.
It depends upon the actual content of his matrix which can vary.
Rod
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Consumer products can be 90% water. That's OK with us.CPG
He is doing biodiesel waste which should contain a bunch of water.
Rod
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
90% ! That is impressive.
Can you use 5mg or more of sample per mL , and look for levels below 1000ppm with the BSTFA type reagents?
I assume he probably won't be looking at very low levels ( xx ppm) in the biodiesel waste.
best wishes,
Rod
Can you use 5mg or more of sample per mL , and look for levels below 1000ppm with the BSTFA type reagents?
I assume he probably won't be looking at very low levels ( xx ppm) in the biodiesel waste.
best wishes,
Rod
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
About 30 years ago we took on a project for a customer to assay for trace levels of glycerin in bar soaps, for legal claims. These were bars made by continuous process instead of the kettle process, so only trace levels remained. We could readily quantitate levels like 0.01% in the product.
High glycerin levels in transparent bar soaps could be about 10%.
High glycerin levels in transparent bar soaps could be about 10%.
-
- Posts: 264
- Joined: Sun Nov 14, 2010 2:20 pm
What type of chromatography/detection makes the most sense for you depends on the matrix it is in and on what else you want to measure along with
I use HILIC on the Water's amide... For HPLC with RI detection, for UPLC with ELSD. That makes the most sense for what we do. Samples are aqueous. We are looking for 1 g/L and up
- Karen
I use HILIC on the Water's amide... For HPLC with RI detection, for UPLC with ELSD. That makes the most sense for what we do. Samples are aqueous. We are looking for 1 g/L and up
- Karen
-
- Posts: 775
- Joined: Mon Dec 01, 2008 9:59 pm
here is highly sensitive UV-VIS method for analysis of glycerol:
http://www.sielc.com/Compound-Glycerol.html
Contact me if you have questions.
http://www.sielc.com/Compound-Glycerol.html
Contact me if you have questions.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
9 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 22 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science