Repeated Sample RSD failures

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Using a very robust (and simple) HPLC method that works constantly on other machines on a Waters Alliance HPLC I keep getting erroneous injections or typically high sample RSD's. The system has been OQ'd and PM'd very recently and passed all the basic points on this but still suffers from these failures. There is a new lamp fiitted too. Occasionally I experience the "plunger over pressure" error code but would this have any bearing?

I am a novice.....please help!!
Plunger over pressure error is certainly worth investigating.
Without more information about your samples and solutions in use, there's not much more we can offer.
Thanks,
DR
Image
Plunger over pressure error is certainly worth investigating.
Always fix first what you KNOW is bad.

See above.
What CPG said...Did you PQ the system post PM? A plunger overpressure error definitely isn't right.
http://the-ghetto-chromatographer.blogspot.com/
What is a "high" RSD? Waters injectors aren't as precise as others
Where can I buy the kit they use in CSI?
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 3 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry