Detection Limit for methanol content using headspace GC
Posted: Fri Dec 07, 2012 8:59 pm
by Shalvin
Hi guys
I am a MSc student and am confused about how to actually work out the detection limit...to get the signal to noise ratio value.
I am working on biodiesel.
I really need some help on this.
Shalvin
Re: Detection Limit for methanol content using headspace GC
Posted: Mon Dec 10, 2012 10:12 am
by Johnny Rod
What is your question? The chromatography data system can calculate the signal:noise ratio (peak height divided by noise value) or you can work it out with a ruler if you are using a chart recorder. Inject a sample of known concentration, scale the peak down to a signal:noise of 3, this is your LOD concentration.
Re: Detection Limit for methanol content using headspace GC
Posted: Tue Dec 18, 2012 7:02 pm
by chromatographer1
There are two detection limits in play here.
One is the detection limit of the detector, calculated by the response of a known amount placed onto the column to the 'noise' of the detector, and
a second LOD which is how much can you detect by headspace recovery from the biodiesel sample using a certain set of parameters.
Both are determined experimentally with prepared standards of a range of concentrations to where no methanol is detected.
They are only calculated after collecting data and evaluating it. Most analysts prefer to assign a limit where the peak response is at least 3 times the amount of detector noise.
Best wishes,
Rod
Re: Detection Limit for methanol content using headspace GC
Posted: Tue Dec 18, 2012 7:54 pm
by Shalvin
Hi Rod
Thanks.
Yes, I am working in that line. Hopefully I can work it out.
Shalvin