My test runs have two problems: loading discrimination and ghost peaks. After further testing, I determined that the liner and/or the inlet need to be cleaned/changed.
I am here seeking opinions on selecting a liner for this assay. what is the best liner technically? My main concern is the high boiling point component discrimination and contamination.
Currently, the GC has a HP/Agilent liner with cup design, filled with 10% OV-1 on Chromosorb, and also a bit glass wool on top, as shown here at http://www.sigmaaldrich.com/catalog/pro ... ®ion=US.
I will be most interested in FAME compositions from C16 to C26, and mainly the heavy ones.
Column: OmegaWax (Polyethylene Glycol coated) 30m x 0.25mm x 0.25um
Sample: FAME in hexane ( the organic liquid phase from a direct transesterification tube)
Injection vol.: 0.5 micro liter manually.
Carrier gas: helium at 2 ml/min @ 80C.
Injector temperature: 250 C
FID detector temperature: 260C
Split flowrate: 35 ml/min @ room temp
Purge flowrate: 3 ml/min@ room temp
Air flowrate: 415 ml/min@ room temp
Hydrogen fule rate: 30 min/min@ room temp
Nitrogen makeup flowrate: 38 min/min@ room temp
Temp: 80C, 2min --> 40C/min to 170C --> hold at 170C for 5 min --> 2 C/min to 220C --> hold at 220C for 25 min.
I searched the forum, and found this old post mentioned a FocusLiner liner at posting.php?mode=reply&f=2&t=8494.
Also, this article recommends http://www.sepscience.com/Techniques/GC ... rimination. Recommend a tapered liner.
Is a tapered liner lined with glass wool a good choice for FAME assay?
How often do you change your inlet liner? I understand the change frequency could be affected by the particular samples, but I don't have any idea how often they need to be changed. Weekly? Monthly? Or roughly how many injections?