Hey Guys, just a question for people with experience running the agilent 5973 mass spec. This question came up at my lab and I am curious what everyone's opinion is. How often would you expect to run the Mass Axis calibration routine on a 5973? is it, as needed, daily... etc.

thanks!
I would do it every time I needed to vent the GC/MS.
If you run a tune check or check standard before running samples, I would do it if the mass assignments are off or if there's a sensitivity change. If you don't run a tune check or check standard before running samples, I would do it before each run. I would do as yama001 suggested, when I did any maintenance, or if I changed filaments. Always check the air/H2O to make sure you're leak free before doing anything else. Good luck.
I run mass axis calibration and peak width adjustment as the final step of any manual tune I do.
4 posts Page 1 of 1

Who is online

In total there are 118 users online :: 0 registered, 0 hidden and 118 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 118 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry