Page 1 of 2
negative intercept in calibration curve
Posted: Wed Dec 05, 2012 11:57 am
by kaka_munda
Hi everybody,
I am having negative intercepts for some of my analytes I injected on GC-uECD. I am wondering if I should force the origin to zero or just add the absolute value of intercept in the calculation?
regards,
kaka
Re: negative intercept in calibration curve
Posted: Wed Dec 05, 2012 12:17 pm
by Don_Hilton
More details are required. The negative intercept can indicate real things going on in the chromatographic system - in which case it is telling you things that can not be ignored - or it may be the result of noise - in which case the better option is reconsideration of calibration levels and number of injections at those levels. (Or the difference actually is so small that deviation from zero can be ignored.)
How are you making up standards and in what matrix? What kinds of compounds are you running? Are you using an internal standard?
Posted: Wed Dec 05, 2012 12:32 pm
by kaka_munda
Thanks for your reply. Standards are hexachlorobenzene, PCB101, penta/hexa-chloroethanes and mirex and they are in isooctane. I made 5 levels by serial dilution spanning three order of magnitude. The intercept value is significant and not near to zero.
Posted: Wed Dec 05, 2012 12:40 pm
by Peter Apps
Hi Kaka
An ECD is not quite linear in response, if you are calibrating over a wide range, or using high concentrations and then extrapolating below the range for which you have actual measurements this might be why you get a negative intercept.
Posted: Wed Dec 05, 2012 1:00 pm
by kaka_munda
Hi Kaka
Hi Peter,
I see your point but the regression among different levels is pretty linear (R2= 0.99) and my sample points fall between the calibration levels. So, in this case, should I ignore intercept?
best,
Kaka
Posted: Wed Dec 05, 2012 1:21 pm
by Peter Apps
Hi Kaka
Just to be certain - you are talking about the intercept on the Y axis ? So you have a positive intercept on the X-axis.
What is the calibration equation ?
Posted: Wed Dec 05, 2012 1:54 pm
by kaka_munda
Hi Kaka
Yes, it is Y intercept (value is - 2E+09). The equation for pentachloroethane is:
y=(3E+11)x - 2E+09
Kaka
Posted: Wed Dec 05, 2012 2:04 pm
by Peter Apps
So the intercept is 2/3 of a percent of the slope.
What are x and y for your lowest calibration point ?
Posted: Wed Dec 05, 2012 2:23 pm
by kaka_munda
For lowest point, x is 3.68E-02 and y is 8.4E+09.
Posted: Wed Dec 05, 2012 2:59 pm
by Peter Apps
The intercept is too big in comparison to your data points to just ignore it. You need to trouble-shoot the analysis; either your calibration standards have lower concentrations than you think that they have, or you have a constant bias in your instrumental response.
Posted: Wed Dec 05, 2012 3:07 pm
by kaka_munda
Thanks but I got negative intercept with this compound only and not for others. I injected them as a mix with equal concentrations.
Posted: Wed Dec 05, 2012 3:45 pm
by Peter Apps
If the other compounds are OK then there is probably nothing wrong with the instrument. How do you make up you standards ?, at what stage do you have all the compounds in solution together ?
Posted: Wed Dec 05, 2012 5:35 pm
by kaka_munda
right from the beginning. I made a stock solution containing the mix of 5 components and then the stock solution was serially diluted to 5 levels.
Posted: Thu Dec 06, 2012 6:54 am
by Peter Apps
Step one is to make the stock solution again - by far the most likely explanation is operator errror at that stage.
Posted: Thu Dec 06, 2012 12:26 pm
by Don_Hilton
Do you use an internal standard? And if so, the same internal standard for all compounds? And does an injection of solvents with no added internal standard or analyte give any peak - particularly in the location for the internal standard?