by
Yama001 » Fri Nov 30, 2012 12:55 pm
Hmm - that is a pretty long run time; perhaps it is needed to get the C17/Pristane and C18/Phytane separation? (BTW, it seems that separation is easier to achieve with a 0.32 mm ID column rather than a 0.25 - annoying. An MS rather than an FID seems to make it harder as well - likely the vacuum at the outlet screws things up a bit).
I currently use the following: RTX-5 column, 30m X 0.32mm X 0.25um film. Hydrogen carrier at 6.2 cc/min constant flow (sometimes when the C19/OTP gets bad dropping this to ~ 4 cc/min helps). FID make up is 10 or 20 cc/min of Helium.
2uL split injection 3:1 (the crappy 6890/7890 inlets seem to have trouble with splitless and early eluters, otherwise I would have stuck with splitless).
Temperature program: 0.5 min at 60, 30 deg/min to 315, hold 3.5 min. These conditions give me no trouble with C19/OTP, but back when I needed the pristane/phytane stuff routinely, I believe I needed to slow that down to maybe a 25 min run.
I can get a year or more out of a column unless I get a lot of clean sediments that deposit a lot of solid material - that seems to be the worst matrix, since I cannot dilute out the garbage. The C19/OTP is never a complete baseline separation; the valley height is around 2 to 5 percent max between them. I tend to change the column due to baseline rise issues rather than separation concerns.
Hope that helps out.