OTP shift

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Anbody have any ideas on how to prevent o-terphenyl from shifting into C19 on an FID alkane run?

Generally when this happens I install a new RTX-5 column, but I know there has got to be some steric hindrance going on and I would like to get more time out of my columns without significant changes to my aquisition method. I'm already getting 6 mos ~ 2000 injections, but I'm wondering if I can't do better.

Alkanes 8-40, pristane, phytane. OTP surrogate and 5 alpha androstane as a quantifying IS. ~69 min run.
Hmm - that is a pretty long run time; perhaps it is needed to get the C17/Pristane and C18/Phytane separation? (BTW, it seems that separation is easier to achieve with a 0.32 mm ID column rather than a 0.25 - annoying. An MS rather than an FID seems to make it harder as well - likely the vacuum at the outlet screws things up a bit).

I currently use the following: RTX-5 column, 30m X 0.32mm X 0.25um film. Hydrogen carrier at 6.2 cc/min constant flow (sometimes when the C19/OTP gets bad dropping this to ~ 4 cc/min helps). FID make up is 10 or 20 cc/min of Helium.

2uL split injection 3:1 (the crappy 6890/7890 inlets seem to have trouble with splitless and early eluters, otherwise I would have stuck with splitless).

Temperature program: 0.5 min at 60, 30 deg/min to 315, hold 3.5 min. These conditions give me no trouble with C19/OTP, but back when I needed the pristane/phytane stuff routinely, I believe I needed to slow that down to maybe a 25 min run.

I can get a year or more out of a column unless I get a lot of clean sediments that deposit a lot of solid material - that seems to be the worst matrix, since I cannot dilute out the garbage. The C19/OTP is never a complete baseline separation; the valley height is around 2 to 5 percent max between them. I tend to change the column due to baseline rise issues rather than separation concerns.

Hope that helps out.
Thanks for the reply Yama. I do have to separate pristane/phytane from the alkanes, and unfortunately, I can't break legacy with my aquisition method. I've got a significantly shorter run, but I can't use it for the current flywheel project.

Generally I can separate the C19/OTP as you said, but when OTP begins shifting, I switch out my columns, and I was wondering if anybody had any magic tricks to lengthen column life.

I appreciate the input.
3 posts Page 1 of 1

Who is online

In total there are 36 users online :: 1 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry