Repeated Sample RSD failures
Posted: Tue Dec 04, 2012 2:34 pm
by djones82
Using a very robust (and simple) HPLC method that works constantly on other machines on a Waters Alliance HPLC I keep getting erroneous injections or typically high sample RSD's. The system has been OQ'd and PM'd very recently and passed all the basic points on this but still suffers from these failures. There is a new lamp fiitted too. Occasionally I experience the "plunger over pressure" error code but would this have any bearing?
I am a novice.....please help!!
Re: Repeated Sample RSD failures
Posted: Tue Dec 04, 2012 3:25 pm
by DR
Plunger over pressure error is certainly worth investigating.
Without more information about your samples and solutions in use, there's not much more we can offer.
Posted: Tue Dec 04, 2012 3:27 pm
by Consumer Products Guy
Always fix first what you KNOW is bad.
See above.
Posted: Wed Dec 05, 2012 3:37 pm
by juddc
What CPG said...Did you PQ the system post PM? A plunger overpressure error definitely isn't right.
Posted: Mon Dec 10, 2012 10:26 am
by Johnny Rod
What is a "high" RSD? Waters injectors aren't as precise as others