Hi,
Can use the Dean Switch in order to connect two GC columns consequently?
Thanks
Hi,
Can use the Dean Switch in order to connect two GC columns consequently?
Thanks
If you want to connect the end of one column to the beginnning of the other, without adding any gas at the connection or diverting any of the carrier gas or sample to a different flow path, then a Deans switch is not what you need. You can use either a press-tight type of connector, or a metal capillary connector.

Peter
Peter Apps
Of course, you can attach the two columns into a single train with the Deans switch with the accompanying dead volume. It would be a very expensive connector though. But I guess it beats spending money if you already have one.

Peter's comments are correct of course.

Rod
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry