- Posts: 21
- Joined: Tue Jun 05, 2012 5:07 am
last week I did my first maintenance on a GC/MS. I cleaned the source, changed both filaments (trimmed Column at the GC, changed Liner and so on ).
After pumping down the MSD and starting it all I did my first Tune after maintenance yesterday.
EM volts were about 1600 (before that the maintenance, they raised over half a year from 1400 to 1700) Repeller was 15,05.
The Values for water (10%) nitrogen (17%) oxygen(5%) an Co2(0,7%) were quite high but felt down with every tune. So I thought, this is maybe from venting the MSD.
When I tried to start a run I didn’t get any signal. So I thought I might have overtighten the column Nut to the MSD.
I loosed the nut a little bit and got signal in the run I did after.
When I did a Tune after this first run I got huge values for Water/Nirogen/Oxigen (>1000%) so I obviously produced a leek with loosing the nut. After tighten the nut again the values for Water/Nirogen/Oxigen looked way better (4%/7%/2%) but the EMVolts and the Repeller went up a lot.
The EMVolts is now 2275 and the Repeller 23,25.(on both Filaments)
What does this rise mean?
I read that you should change the Electron Multiplier when reaching the about 2200 Volt. Is this true?
How can I reduce the optimum for the EMVolts. Is it just by cleaning the source (which I have just done) or are there other procedures?
What a normal values for water, nitrogen and oxygen? Less than 3 %? or is less than 5% okay (which I had in The past)?
Thank you!