Coffo » Thu Nov 29, 2012 9:27 am
Usually 1200 series LCs have an initial IP address of 192.168.254.11 ...
In order to connect you should assign a similar domain IP address to the network card you are using to connect to the LC (i.e. 192.168.254.xxx, where xxx can be any number from 1 to 255 except 11).
If you are running Windows 7, this can be done by going to Control Panel > Network and Internet > Network and Sharing Center. If you have two network cards, select the one connected to the HPLC (this will be labelled "No internet access"). When you click the Local Area Connection link, click on Properties, highlight "Internet Protocol Version 4 (TCP/IPv4)" and click on Properties. Now select "Use the following IP address" and enter 192.168.254.100 (for example - as long as the first three sets of digits are the same as those for the instrument and the four set is different) and 255.255.255.0 as the subnet mask and click OK.
You should now be able to communicate with the HPLC through this network card and still connect to your company share drive through the other network card (it will normally have an automatically assigned IP address).
[If you have already assigned another IP address to the instrument when you first installed ChemStation and ran the Add Instrument utility, you will need to change the above IP address to this domain or else run the Add Instrument program again and manually change the instrument's IP address back to 192.168.254.11].
The installation manual (p26-28) gives instructions for Windows XP and also suggests using 10.1.1.100 for the computer and 10.1.1.102 for the HPLC addresses.