They're fine for something undemanding like analysis of a drug or other small molecule, which describes a large fraction of HPLC analyses. They are not fine for HILIC of peptide digests, intact membrane proteins, etc. They are also inferior to materials with thick, well-hydrated coatings in cases where an analyte is poorly-retained in HILIC or is only marginally soluble in the mobile phases needed for retention (e.g., an oligosaccharide in 80% ACN). The better hydrated a stationary phase material is, the less organic solvent is necessary to get a given degree of retention of an analyte in the HILIC mode.
Silica columns are cheap and simple, and can be more durable under some conditions since they have no coating to lose. If they work adequately for a particular application, then by all means go ahead and use them.