mobile phase switch from reverse phase to normal phase
Posted: Thu Nov 29, 2012 12:38 am
by jiangds06
If mobile phase does not switch well, will it cause clogging?
for example from water/ACN to iPrOH/Hexane. If it does not switch well, then four solvent will present in the column simultaneously. But still, I do not think it will clog the column.
Re: mobile phase switch from reverse phase to normal phase
Posted: Thu Nov 29, 2012 8:00 am
by prepcolumn
yes it does not clog because you don't have any buffer to cause clogging in your example

. But if you have buffers in your RP mobile phase then it clogs.
But the problem switching from NP to RP or vice versa is the miscibility of the solvents you are using.
Re: mobile phase switch from reverse phase to normal phase
Posted: Thu Nov 29, 2012 2:53 pm
by jiangds06
So it only cause UV detection and pressure fluctuation problem, right?
yes it does not clog because you don't have any buffer to cause clogging in your example

. But if you have buffers in your RP mobile phase then it clogs.
But the problem switching from NP to RP or vice versa is the miscibility of the solvents you are using.
Re: mobile phase switch from reverse phase to normal phase
Posted: Thu Nov 29, 2012 3:00 pm
by Consumer Products Guy
yes it does not clog because you don't have any buffer to cause clogging in your example

. But if you have buffers in your RP mobile phase then it clogs.
But the problem switching from NP to RP or vice versa is the miscibility of the solvents you are using.
Correct. Typically one would flush everything with IPA (isopropyl alcohol) as the interim step between RP and NP, and between NP and RP.
Re: mobile phase switch from reverse phase to normal phase
Posted: Thu Nov 29, 2012 8:34 pm
by jiangds06
If it is not properly switched, what will happen?
yes it does not clog because you don't have any buffer to cause clogging in your example

. But if you have buffers in your RP mobile phase then it clogs.
But the problem switching from NP to RP or vice versa is the miscibility of the solvents you are using.
Correct. Typically one would flush everything with IPA (isopropyl alcohol) as the interim step between RP and NP, and between NP and RP.
Re: mobile phase switch from reverse phase to normal phase
Posted: Fri Nov 30, 2012 3:37 pm
by prepcolumn
Simply; you cannot control any of your chromatographic and instrument parameters because of immiscible phase formation..
Re: mobile phase switch from reverse phase to normal phase
Posted: Fri Nov 30, 2012 6:02 pm
by jiangds06
will a overnight IPA flushing restore the column performance.
Or column can be permanently damaged by emulsion formation in the column?
Simply; you cannot control any of your chromatographic and instrument parameters because of immiscible phase formation..
Re: mobile phase switch from reverse phase to normal phase
Posted: Fri Nov 30, 2012 10:20 pm
by prepcolumn
just try and wish to be lucky..
Re: mobile phase switch from reverse phase to normal phase
Posted: Sat Dec 01, 2012 4:37 pm
by uzman
For reverse phase columns , even pure hexan causes no significant problem , simply you may wash the column with Tetrahydrofuran , then water and return back to your mobile phase.
( Long term use of Tetrahydrofuran may cause hardening of PEEK materials and may burst them )
For normal phase columns , mostly for silica , water and alcohols cause deactivation of column , so you can not get separation.
According to my experience , the most effective way to recover a silica column is washing with pure ethyl acetate for half an hour at 1ml/min flow rate.
Then wash it with hexane for another half an hour and return back to your NP mobile phase.
Good luck.
Re: mobile phase switch from reverse phase to normal phase
Posted: Mon Dec 03, 2012 2:40 pm
by jiangds06
Thanks a lot.
How about Polysaccharide Chiral HPLC Columns? Does emulsion formation damage this phase?
For reverse phase columns , even pure hexan causes no significant problem , simply you may wash the column with Tetrahydrofuran , then water and return back to your mobile phase.
( Long term use of Tetrahydrofuran may cause hardening of PEEK materials and may burst them )
For normal phase columns , mostly for silica , water and alcohols cause deactivation of column , so you can not get separation.
According to my experience , the most effective way to recover a silica column is washing with pure ethyl acetate for half an hour at 1ml/min flow rate.
Then wash it with hexane for another half an hour and return back to your NP mobile phase.
Good luck.
Re: mobile phase switch from reverse phase to normal phase
Posted: Mon Dec 03, 2012 8:02 pm
by tom jupille
Those are a different animal because the stationary phase is not (usually) bonded to the substrate. That means that residual water can actually damage the packing material.
Re: mobile phase switch from reverse phase to normal phase
Posted: Sat Dec 22, 2012 10:37 am
by jiangds06
this column is also can be used in Reverse phase, so probably this is a different animal you mentioned.
Will switching from NP to RP damage the column?
Those are a different animal because the stationary phase is not (usually) bonded to the substrate. That means that residual water can actually damage the packing material.
Re: mobile phase switch from reverse phase to normal phase
Posted: Sat Dec 22, 2012 5:53 pm
by tom jupille
Will switching from NP to RP damage the column?
The best answer to that would come from the column manufacturer.