I am using a system with a multi-mode inlet and blackflush. We are using a cold splitless injection.
My problem is more than half of the calibration curves are quadratic. This would be fine, but those compounds with quadratic curves are showing inconsistent response througout the length of an average run. I can't seem to get a calibration check standard to pass within 20%.
I tried running a sequence with 5 repetitions of: standard, followed by 3 samples. For the quadratic curve compounds, the response for the standard started poor, got better, but then got poor again. I think this is normal, but not to the extent that the injections varied. Some injections of the same standard were even 40% different for those trouble compounds. For the linear ones, the % difference was within 20% without fail.
Has anyone had issues with this and identified the root of the problem? If not, any recommendations on how to remedy this? When Agilent ran some demo standards for us before we bought the instrument, they had the same problem with quadratic curves but told us it was something we could fix in our method. I'm starting to wonder if that's really true.
Thanks so much!
-Chrissy