Agilent FLD test in Empower

Posted: Tue Nov 27, 2012 6:28 pm
by mburk
I have an Agilent 1100 G1321A fluorescence detector that needs to be calibrated. We can run the calibration using the G4208 controller but we can't check the raman band of water using the controller. The detector manual describes how to do it using Chemstation or using the old 1200 style controller. We don't have an 1200 style controller and we are using Empower 2 instead of Chemstation. Does anyone know if this test can be done through Empower?

Re: Agilent FLD test in Empower

Posted: Fri Nov 30, 2012 12:59 am
by Bull76
The test can be done in empower as Waters are qualifing agilent systems. They have an IM and test data for the raman gain of water for the FLR detector.
Depends on where you are as to how easy this can be.