lmh » Fri Jul 13, 2012 10:24 am
You will always get a leak back out of the syringe if you get "normal" Hamilton-style syringes with a simple metal plunger. You absolutely must, must, must get gas-tight ones (recognisable by a white plastic tip on the end of their plunger).
The reason I say this so strongly is that if you are teeing together 300uL/min from the LC pump and 5uL/min from the syringe, and combining these to flow into the spray chamber, then even the tiniest leak in the syringe will mean that Absolutely No analyte will get to the mass spec at all! You only need a back-leak of 5uL/min, which is barely noticeable, to give you no signal except solvent. Even the gas-tight syringes do not last for ever. When they get loose, it's time to replace them, and false economy to try to make do with an old leaky one.
What solvent signal are you seeing? If you have specific ions, like 279 and 391 in positive mode, then you do have dirty solvents or tubing. If it is a general fuzzy mess through the whole range of the spectrum, there may be nothing wrong. If you have a ladder of peaks, it might be a salt contamination from somewhere, but I'd be very surprised at this.
If you're getting no signal when you do an injection from the autosampler in the absence of any column, then there are a few other things to check. If this is a Surveyor autosampler then it has a double-diameter syringe thing with a tiny inner plunger (which makes the injection) and a wider outer plunger (which does the high-volume washes). The only thing that decides whether the motor drives the inner, narrow plunger, or the outer, wide plunger, is how far up or down it is. If there is a misunderstanding between the software and the syringe about how far up or down it is, then it will be using the wrong plunger, and it may squirt the sample right through the sample loop, and to waste. Always, Always before you start a sequence, purge the syringe, make sure it's full of liquid (a few small bubbles are acceptable if they stay the same through purging), and then home it (in the direct control menu of Xcalibur). If it makes grinding noises, it was in the wrong place. If this happens often, clean the lead-screw that drives the syringe, and check the needle isn't blocked.
Also do check that the tubing from injection port to injection valve is original and hasn't been cut/shortened/modified; its volume is critical and should match a value in the Configuration set-up.
You can also check the needle isn't blocked (especially if the syringe makes grindy noises, see above). Use Direct Control to get the needle to its removal position, locate the little black plastic bit at the top of the needle and turn sideways to unlock, and draw it out upwards, carefully. Now do a needle wash from direct control and make sure the liquid comes out as a free steady stream. Remember when you do this that the instrument will move the needle position to the wash-station, so be prepared for it to whizz around. If you've freed the needle tubing you will be OK! To do this, you may also need to disable the bit in configuration about sensing whether the door is closed, or fool it into thinking the door is closed by putting a small magnet at the top - but Be Careful not to get injured by moving parts. Do this at own risk, with sense! Good luck!