Hi Guys!

I have a question. After the analysis of residual solvents in DMF, doing analysis of residual solvents in the water. how to completely clean the headspace system from DMF? (DMF- N,N-Dimethylfromamide (Sorry)

GC-HS agilent 7890A-G1888


many thanks
Greetings!

Your headspace system it's automated? The analysis it's from gc/ms or gc/fid?

It's dimethylsulfoxide or dimethlyformamide ? Dimethylsulfoxide is commonly expressed by DMSO. Either way i think the procedure will be the same for both.

If it's automated , remove the syringe, rinse the syringe about 3 times with acetone and 3 times with methanol. I reccomend a five minutes ultrassonic bath in methanol too, if you have this equipment of course. Put the syringe in the automated headspace and purge it for 5 minutes with nitrogen or other (cheap) gas that you use, run a blank.

if it's not automated, rinse with the solvents listed above about 5 times each. Wait the syringe to get dry and run a blank.

If in both cases you see something indicating contamination, consider change the septum and liner too.

:D

t.b.
T.B.
Greetings!

Your headspace system it's automated? The analysis it's from gc/ms or gc/fid?

It's dimethylsulfoxide or dimethlyformamide ? N,N-Dimethylsulfoxide is commonly expressed by DMSO. Either way i think the procedure will be the same for both.

If it's automated , remove the syringe, rinse the syringe about 3 times with acetone and 3 times with methanol. I reccomend a five minutes ultrassonic bath in methanol too, if you have this equipment of course. Put the syringe in the automated headspace and purge it for 5 minutes with nitrogen or other (cheap) gas that you use, run a blank.

if it's not automated, rinse with the solvents listed above about 5 times each. Wait the syringe to get dry and run a blank.

If in both cases you see something indicating contamination, consider change the septum and liner too.

:D

t.b.


Ops!

your equipment is this : http://www.chem.agilent.com/Library/bro ... 6491EN.pdf
right?

I have a CTC-pal system coupled with a 5975C-7890A Agilent, an automated headspace system pretty simple acctually. I don't know if the procedure i listed will be able to clean your system. I really don't know how your system works.

Sorry.... :( ....
T.B.
Ops!

your equipment is this : http://www.chem.agilent.com/Library/bro ... 6491EN.pdf
right?

I have a CTC-pal system coupled with a 5975C-7890A Agilent, an automated headspace system pretty simple acctually. I don't know if the procedure i listed will be able to clean your system. I really don't know how your system works.

Sorry.... :( ....
thanks anyway :)))
DMF is miscible with water, just put a mL of water in 10 vials and then equilibrate them in the oven at like 110 for a few minutes and inject, set the other heated zones to elevated temp and your GC's inlet to split mode with a few hundred mL of split flow. Set the oven above 100 C too, so no water collects anywhere. You could even remove your column if you want, just install an inch of two of column with a column nut into the inlet so it holds pressure.

If you still see DMF peaks after cleaning you may need to run more vials but I would bet it will be gone. The syringe is more difficult to get to in a G1888 than a CTC sampler. There is also significant flow path and a 6 port valve. All of it is heated so a bit of steam should help purge everything out, but I wouldn't go taking anything apart unless I needed to.
DMF is miscible with water, just put a mL of water in 10 vials and then equilibrate them in the oven at like 110 for a few minutes and inject, set the other heated zones to elevated temp and your GC's inlet to split mode with a few hundred mL of split flow. Set the oven above 100 C too, so no water collects anywhere. You could even remove your column if you want, just install an inch of two of column with a column nut into the inlet so it holds pressure.

If you still see DMF peaks after cleaning you may need to run more vials but I would bet it will be gone. The syringe is more difficult to get to in a G1888 than a CTC sampler. There is also significant flow path and a 6 port valve. All of it is heated so a bit u very of steam should help purge everything out, but I wouldn't go taking anything apart unless I needed to.
Thank you very much :))
6 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry