Accela pump pressure problems
Posted: Wed Nov 21, 2012 10:36 pm
by nitrendipin
Hello!
I've got two used accela pumps that should be used as a replacement for jasco 980 pumps in our unit. The problem is that they both show the same pressure fluctuations (about 30 bar) at a flowrate of 200µl/min . I already changed the piston seals, but that didn't solve the problem. I also checked the checkvalves; they seem to be fine.
Does anyone have experience with accela pumps or is experiencing the same symptoms?
Best regards
Re: Accela pump pressure problems
Posted: Thu Nov 22, 2012 6:26 pm
by coffee bean
My impression is that you have to prime the pulse dampeners. If they are dry, they won't function as designed. Check the manual for detailed instructions.
Posted: Fri Nov 23, 2012 8:12 pm
by nitrendipin
I have also considered that and primed the pulse-dampener with isopropanol for approximately 20 minutes. I also tried Methanol. The only thing that I didn't do is have a syringe installed whilst priming. Maybe there should be some backpressure during that process. I will give that a try and post my results here!
Posted: Sat Nov 24, 2012 5:08 pm
by elsico
Was same problem with customer Accela pump. Prime it with water/Acn/water/isopropanol. Not much success, was about 50-60 bar and best result was on isopropanol with about 15. No leaks and no ideas. So my suggestion to take back to jasco pump and add SSI™ LO-Pulse™ damper. Best solution and don't waste your time.
Posted: Wed Nov 28, 2012 6:28 pm
by leadazide
At what pressure are you seeing the 30 bar pressure difference? Sometimes I have succesfully run the pump up to around 800+ bar to "stick everything together" and after that perfomance was improved..
Posted: Tue Dec 11, 2012 12:14 pm
by vogelius
If you have a flow trough the system, try to clean with Water-10% HNO3-water, very god for checkvalve problem what I think it is.
BR